P40 data separation in modelspace

Post by marcogm »

Currently scanning following a method where we would do a coarse Target All scan and after one or several sub-scans of the areas we need with better resolution.

Having an issue where, viewing the modelspace of only one station there is separation in the scans as seen in the images below. One line of data shows the coarse scan and the offset line is the higher-resolution scan. This would happen on most stations, but in other stations the different scans would be aligned.
2.png
1.png
Thoughts?

Thanks!
You do not have the required permissions to view the files attached to this post.
rpfischner
I have made <0 posts
I have made <0 posts
Posts: 7
Joined: Tue Feb 24, 2015 12:18 pm
9
Full Name: Robert Pfischner
Company Details: Feldman Land Surveyors
Company Position Title: BIM Manager
Country: USA
Linkedin Profile: No
Been thanked: 1 time

Re: P40 data separation in modelspace

Post by rpfischner »

I've seen this when the scanner is set up on wood flooring or roofing membrane. When you run the coarse scan you've generally walked away a few feet or more, but during the fine scan you may be standing next to the scanner and the weight of your body may slightly impact the position of the tripod.

Pretty rare but I've definitely seen it on softer surfaces.
JNally
I have made 90-100 posts
I have made 90-100 posts
Posts: 98
Joined: Mon Feb 01, 2016 12:00 pm
8
Full Name: Joe Nally
Company Details: EPS
Company Position Title: Lead Surveyor
Country: Ireland
Linkedin Profile: Yes
Location: IRL
Has thanked: 13 times
Been thanked: 18 times

Re: P40 data separation in modelspace

Post by JNally »

Id agree with Robert here.

Have seen this before with change of height from walking away.
brucewhitesterling
I have made 10-20 posts
I have made 10-20 posts
Posts: 12
Joined: Wed Feb 23, 2022 6:36 pm
2
Full Name: Bruce White
Company Details: Sterling Systems
Company Position Title: President
Country: United States
Linkedin Profile: Yes
Been thanked: 3 times

Re: P40 data separation in modelspace

Post by brucewhitesterling »

Yes, I believe the above answers are correct, however, you might also just use the "Check and Adjust" calibration test just to make sure that you are still in tolerance.
marcogm
I have made <0 posts
I have made <0 posts
Posts: 2
Joined: Wed Jun 01, 2022 2:50 pm
1
Full Name: Marco G
Company Details: Benchmark
Company Position Title: Drafting
Country: Canada
Linkedin Profile: No

Re: P40 data separation in modelspace

Post by marcogm »

We did the "Check and Adjust" calibration test and it did not change the results.

I have also seen that separation while being on unstable ground (roofing membrane, etc). However, I have been seeing the separation mentioned in this topic while the tripod being set up the tripod in stable ground (pavement).

The hypothesis at the moment is that it may be caused by the scanner's firmware being out-of-date. We are in the process of updating the firmware hopefully later this week.

Thanks for the replies! Will let you guys know if the update made the difference.